Replacing the Patient Switch did not do the magic I was hoping for. Back to square one.

2
Thanks James,

much appreciated. This was not meant to be a service request. Sorry if I gave that impression. I know the age of the machine is 15 years, but that doesn't mean I won't try to fix it.

I thought this is a forum for Medmont users where I could just post the problem we are experiencing in the hope that someone has had something similar.

For those who are interested, I have now tried to use a powered USB connection, but that didn't do the trick unfortunately. Will try to replace the patient switch. Maybe that one is the culprit.

3
So it is an M700(Serial Number M434130) and I am running it with Studio 6. And yes, the switch presses are correctly detected and recorded by the software. The only thing I can now think of is that we changed to a Win10 machine about half a year ago which doesn't have a serial port or VGA port anymore, so we are running the M700 through a USB port. Maybe not enough power? Was thinking of adding a powered USB hub to the laptop maybe? Best regards, Reiner

4
Issues/Problems / M700 Background Lighting dims slightly when clicking
« on: October 15, 2019, 01:06:49 pm »
I am not sure if it has always been like this, but it seems that the background lighting is slightly dimming every time one clicks the button. Have checked all connections and they seem ok. Any suggestions what else to test? Would the test results be influenced by this?